Pressure Sensor MPXM2051GST1 spec
Adam_WM
On revision 5 of the MPXM2051G specification, 22 April 2021, on page 5 the Linearity in Table 4 shows a min of -0.3 and a max of +3.0 %Vfss. The maximum of +3.0 % seems high. Is this spec correct?
JozefKozon
Dear Adam,
the application engineer confirmed, that the max. linearity value is a typo and it should be 0.3%VFSS, not the not 3%VFSS.
With Best Regards,
Jozef
Adam_WM
Jozef,
Thanks for the reply. Sorry to be a bother, but on Rev 4 of the MPXM2051G specification the Linearity is listed as -0.3% to +0.3%, page 4, table 3. This is also the case for revision 1, 2, and 3 of the MPXM2051G spec. Revision 4 of the spec is attached.
The revision history on the rev 5 specification also does not mention a change to the Linearity spec.
Thanks for your help with this.
Adam M
JozefKozon
Dear Adam,
thank you for the research you have conducted. Due to the new information you shared I think yes, the max. linearity value is most probably 0.3%VFSS, not 3%VFSS. I have contacted an application engineer for confirming the correct value and asked to fix the mistake in the datasheet rev. 5 if the 0.3% is correct. I will let you know the answer from the application engineer.
With Best Regards,
Jozef
JozefKozon
Dear Adam,
With Best Regards,
Jozef
JozefKozon
Dear Adam,
unfortunately yes. The MPXM2051GS has such high linearity error. If you want a similar gauge sensor with much better linearity error, you can use for example MPX2050, which has the linearity error +/-0.25%.
Please check this link for other differential and gauge pressure sensors up to 115kPa.
With Best Regards,
Jozef
%3CLINGO-SUB%20id%3D%22lingo-sub-1790503%22%20slang%3D%22en-US%22%20mode%3D%22CREATE%22%3EPressure%20Sensor%20MPXM2051GST1%20spec%3C%2FLINGO-SUB%3E%3CLINGO-BODY%20id%3D%22lingo-body-1790503%22%20slang%3D%22en-US%22%20mode%3D%22CREATE%22%3E%3CP%3EOn%20revision%205%20of%20the%20MPXM2051G%20specification%2C%2022%20April%202021%2C%20on%20page%205%20the%20Linearity%20in%20Table%204%20shows%20a%20min%20of%20-0.3%20and%20a%20max%20of%20%2B3.0%20%25Vfss.%20The%20maximum%20of%20%2B3.0%20%25%20seems%20high.%20Is%20this%20spec%20correct%3F%3C%2FP%3E%3C%2FLINGO-BODY%3E%3CLINGO-SUB%20id%3D%22lingo-sub-1792751%22%20slang%3D%22en-US%22%20mode%3D%22CREATE%22%20translate%3D%22no%22%3ERe%3A%20Pressure%20Sensor%20MPXM2051GST1%20spec%3C%2FLINGO-SUB%3E%3CLINGO-BODY%20id%3D%22lingo-body-1792751%22%20slang%3D%22en-US%22%20mode%3D%22CREATE%22%3E%3CP%3EDear%20Adam%2C%3C%2FP%3E%0A%3CP%3Ethe%20application%20engineer%20confirmed%2C%20that%20the%20max.%20linearity%20value%20is%20a%20typo%20and%20it%20should%20be%200.3%25VFSS%2C%20not%20the%20%3CSPAN%3Enot%203%25VFSS.%26nbsp%3B%3C%2FSPAN%3E%3C%2FP%3E%0A%3CP%3EWith%20Best%20Regards%2C%3C%2FP%3E%0A%3CP%3EJozef%3C%2FP%3E%3C%2FLINGO-BODY%3E%3CLINGO-SUB%20id%3D%22lingo-sub-1791612%22%20slang%3D%22en-US%22%20mode%3D%22CREATE%22%20translate%3D%22no%22%3ERe%3A%20Pressure%20Sensor%20MPXM2051GST1%20spec%3C%2FLINGO-SUB%3E%3CLINGO-BODY%20id%3D%22lingo-body-1791612%22%20slang%3D%22en-US%22%20mode%3D%22CREATE%22%3E%3CP%3EDear%20Adam%2C%3C%2FP%3E%3CP%3Ethank%20you%20for%20the%20research%20you%20have%20conducted.%20Due%20to%20the%20new%20information%20you%20shared%20I%20think%20yes%2C%20the%20max.%20linearity%20value%20is%20most%20probably%200.3%25VFSS%2C%20not%203%25VFSS.%20I%20have%20contacted%20an%20application%20engineer%20for%20confirming%20the%20correct%20value%20and%20asked%20to%20fix%20the%20mistake%20in%20the%20datasheet%20rev.%205%20if%20the%200.3%25%20is%20correct.%20I%20will%20let%20you%20know%20the%20answer%20from%20the%20application%20engineer.%3C%2FP%3E%3CP%3EWith%20Best%20Regards%2C%3C%2FP%3E%3CP%3EJozef%3C%2FP%3E%3C%2FLINGO-BODY%3E%3CLINGO-SUB%20id%3D%22lingo-sub-1791194%22%20slang%3D%22en-US%22%20mode%3D%22CREATE%22%20translate%3D%22no%22%3ERe%3A%20Pressure%20Sensor%20MPXM2051GST1%20spec%3C%2FLINGO-SUB%3E%3CLINGO-BODY%20id%3D%22lingo-body-1791194%22%20slang%3D%22en-US%22%20mode%3D%22CREATE%22%3E%3CP%3EJozef%2C%3C%2FP%3E%3CP%3EThanks%20for%20the%20reply.%20Sorry%20to%20be%20a%20bother%2C%20but%20on%20Rev%204%20of%20the%20MPXM2051G%20specification%20the%20Linearity%20is%20listed%20as%20-0.3%25%20to%20%2B0.3%25%2C%20page%204%2C%20table%203.%20This%20is%20also%20the%20case%20for%20revision%201%2C%202%2C%20and%203%20of%20the%20MPXM2051G%20spec.%20Revision%204%20of%20the%20spec%20is%20attached.%3C%2FP%3E%3CP%3EThe%20revision%20history%20on%20the%20rev%205%20specification%20also%20does%20not%20mention%20a%20change%20to%20the%20Linearity%20spec.%26nbsp%3B%3C%2FP%3E%3CP%3EThanks%20for%20your%20help%20with%20this.%3C%2FP%3E%3CP%3EAdam%20M%3C%2FP%3E%3C%2FLINGO-BODY%3E%3CLINGO-SUB%20id%3D%22lingo-sub-1790796%22%20slang%3D%22en-US%22%20mode%3D%22CREATE%22%20translate%3D%22no%22%3ERe%3A%20Pressure%20Sensor%20MPXM2051GST1%20spec%3C%2FLINGO-SUB%3E%3CLINGO-BODY%20id%3D%22lingo-body-1790796%22%20slang%3D%22en-US%22%20mode%3D%22CREATE%22%3E%3CP%3EDear%20Adam%2C%3C%2FP%3E%0A%3CP%3Eunfortunately%20yes.%20The%20MPXM2051GS%20has%20such%20high%20linearity%20error.%20If%20you%20want%20a%20similar%20gauge%20sensor%20with%20much%20better%20linearity%20error%2C%20you%20can%20use%20for%20example%26nbsp%3B%3CA%20href%3D%22https%3A%2F%2Fwww.nxp.com%2Fdocs%2Fen%2Fdata-sheet%2FMPX2050.pdf%22%20target%3D%22_self%22%20rel%3D%22nofollow%20noopener%20noreferrer%22%3EMPX2050%3C%2FA%3E%2C%20which%20has%20the%20linearity%20error%20%2B%2F-0.25%25.%26nbsp%3B%3C%2FP%3E%0A%3CP%3E%3CSPAN%20class%3D%22lia-inline-image-display-wrapper%20lia-image-align-inline%22%20image-alt%3D%22JozefKozon_0-1705560081822.png%22%20style%3D%22width%3A%20400px%3B%22%3E%3Cspan%20class%3D%22lia-inline-image-display-wrapper%22%20image-alt%3D%22JozefKozon_0-1705560081822.png%22%20style%3D%22width%3A%20400px%3B%22%3E%3Cimg%20src%3D%22https%3A%2F%2Fcommunity.nxp.com%2Ft5%2Fimage%2Fserverpage%2Fimage-id%2F258740iD9D595B023104952%2Fimage-size%2Fmedium%3Fv%3Dv2%26amp%3Bpx%3D400%22%20role%3D%22button%22%20title%3D%22JozefKozon_0-1705560081822.png%22%20alt%3D%22JozefKozon_0-1705560081822.png%22%20%2F%3E%3Cspan%20class%3D%22lia-inline-image-caption%22%20onclick%3D%22event.preventDefault()%3B%22%3EJozefKozon_0-1705560081822.png%3C%2Fspan%3E%3C%2Fspan%3E%3C%2FSPAN%3E%3C%2FP%3E%0A%3CP%3EPlease%20check%20this%20%3CA%20href%3D%22https%3A%2F%2Fwww.nxp.com%2Fproducts%2Fproduct-selector%3APRODUCT-SELECTOR%3Fcategory%3Dc265_c231_c82%22%20target%3D%22_self%22%20rel%3D%22nofollow%20noopener%20noreferrer%22%3Elink%3C%2FA%3E%20for%20other%20differential%20and%20gauge%20pressure%20sensors%20up%20to%20115kPa.%26nbsp%3B%3C%2FP%3E%0A%3CP%3EWith%20Best%20Regards%2C%3C%2FP%3E%0A%3CP%3EJozef%3C%2FP%3E%0A%3CBR%20%2F%3E%3C%2FLINGO-BODY%3E