Hi everyone, i am working with blood pressure monitor portable ,

I had finished analog circuit from sensor MPX5050 , then i want to test this circuit , but i don't have no idea , how to generate a blood pressure signal ! ,

can you help me ! Thanks  ,

here is my ciruits

タグ(3)
0 件の賞賛
1 解決策
1,637件の閲覧回数
Joshevelle
Senior Contributor I

Hello a b,

You’d need a blood pressure cuff. The pressure sensor would measure the pressure change in the cuff:

http://flightphysical.com/images/bodyparts/blood-pressure-cuff.jpg

In the above image you can see a blood pressure cuff with a mechanical pressure gauge, instead of that you would connect P1 of the MPX5050.

Please take a look at “Ankle-Brachial Index” image in the following page for detailed information on using a blood pressure cuff:

http://www.ehow.com/how_4425292_use-blood-pressure-cuff.html

If you’re only simulating the circuit, you’d need to generate a signal like the one shown in the following page:

http://www.biopac.ca/Continuous%20Non%20Invasive%20Blood%20Pressure.htm

I’d recommend you to run some test in Matlab or Octave, you can generate such signal by adding sinusoidal signals of different frequencies.

You could also download an artificial biomedical signal generator from the following page from OhioLINK ETD Center:

http://etd.ohiolink.edu/view.cgi/Bhojwani%20Soniya%20Naresh.pdf?akron1193079604

Hope it helps!

Josh

元の投稿で解決策を見る

0 件の賞賛
2 返答(返信)
1,638件の閲覧回数
Joshevelle
Senior Contributor I

Hello a b,

You’d need a blood pressure cuff. The pressure sensor would measure the pressure change in the cuff:

http://flightphysical.com/images/bodyparts/blood-pressure-cuff.jpg

In the above image you can see a blood pressure cuff with a mechanical pressure gauge, instead of that you would connect P1 of the MPX5050.

Please take a look at “Ankle-Brachial Index” image in the following page for detailed information on using a blood pressure cuff:

http://www.ehow.com/how_4425292_use-blood-pressure-cuff.html

If you’re only simulating the circuit, you’d need to generate a signal like the one shown in the following page:

http://www.biopac.ca/Continuous%20Non%20Invasive%20Blood%20Pressure.htm

I’d recommend you to run some test in Matlab or Octave, you can generate such signal by adding sinusoidal signals of different frequencies.

You could also download an artificial biomedical signal generator from the following page from OhioLINK ETD Center:

http://etd.ohiolink.edu/view.cgi/Bhojwani%20Soniya%20Naresh.pdf?akron1193079604

Hope it helps!

Josh

0 件の賞賛
1,637件の閲覧回数
tinhnv92
Contributor I

sorry ; i just want to test in software !

0 件の賞賛