Hi,
The output voltage on the MPX2050 is ratiometric, this means that output voltage of the sensor is directly affected by the voltage supply. Make sure the input supply voltage is constant. If there is a small drop on the input voltage, the output voltage would be affected. For example, if you are using a battery in your system to supply the pressure sensor, there would be a small battery drain that could cause the voltage drop you are seen on the output of the sensor.
Another possible cause is the pressure connected on port P2 of the MPX2050DP, if I understand correctly, you are using the differential (two ported) version of the sensor, and you are only using Port P1 to measure the pressure, while port P2 is unconnected, is my understanding correct?
If it is correct, the local atmospheric pressure can be changing causing a change in the reference of the pressure measurement, which cause a shift on the output voltage of the sensor.
Have a great day,
Jose
-------------------------------------------------------------------------------