MMA8452Q Self Test
MMA8452Q Self Test
2,617 次查看
stanfordhudson
We are using an MMA8452Q and are utilizing the Self Test feature during production test.
The datasheet indicates that when self test is activated, the XYZ axes will be deflected by +45, +64, and +420 LSBs off of the current acceleration acting on the sensor respectively.
Our procedure is as follows:
  Set to ACTIVE mode
  Read X,Y,Z
  Set to ST mode
  Read X,Y,Z
  Compute absolute differences in XYZ for ACTIVE and ST mode.
(We are doing quite a bit of averaging along the way to account for any noise that may exist in the system.)
We routinely get values that are 380 and below for the Z axis.  The XY are usually pretty marginal, but sometimes are dead-on.
Each accel is different.
Can you please give me some insight into why the Z axis is not even close to the +420 specified in the datasheet?
(Note that we do set the accel range to +/-4G.  The Active bit in CTRL_REG1 is cleared before we set ST.  The accel is still during the test).
Stanford Hudson
标记 (3)
8 回复数
1,825 次查看
TomasVaverka
Hello Stanford,
The reason why the Z axis self-test output does not correspond to the typical value specified in the datasheet is that the self-test is not factory trimmed. Generally, the test is considered successful if the minimum magnitude of change (50 counts for all axes at ±4g mode) is achieved.
Tomas
PS: If my answer helps to solve your question, please mark it as "Correct" or “Helpful”. Thank you.
1,825 次查看
danielwerder
Hi Tomas,
I have a question about Self Test range.
About datasheet are XYZ will be deflected by +45, +64 and +420.
Are there no tolerance (min, max) values?
BR,
1,825 次查看
TomasVaverka
Hi Daniel,
We do not specify the limits as the self-test output is not trimmed.
Tomas
1,826 次查看
stanfordhudson
Thank you Tomas for the response.  I am assuming by magnitude you mean the magnitude of the XYZ vector?
(sqrt(x2+y2+z^2)).
Would this then be the correct procedure:
1. Measure xyz in Active mode.
2. Compute magnitude (mag0=sqrt(x2+y2+z^2))
3. Measure xyz in Self Test mode.
4. Compute magnitude (mag1=sqrt(x2+y2+z^2)).
5. Test pass if abs(mag1-mag2)>=50.
Thanks, Stan
1,826 次查看
TomasVaverka
Hi Stan,
Sorry for not being clear enough, the procedure does not require calculating the XYZ vector, just simply compute the difference between the acceleration output of all axes with self-test enabled and disabled as follows:
XST = XST_ON − XST_OFF
YST = YST_ON − YST_OFF
ZST = ZST_ON − ZST_OFF
and if these self-test changes are higher than 50 counts at ±4g mode, the test is considered to pass.
As you are doing correctly, it is recommended to take several samples (e.g. 10) for each axis and average them.
I hope it helps.
Tomas
1,826 次查看
chrislin_wpi
Dear Tomas,
Do you mean each axis (Xst,Yst and Zst) higher than 50 counts at ±4g mode, it's pass?
And how about ±2g mode?
Best Regards
Chris Lin
1,826 次查看
TomasVaverka
Dear Chris,
It is 100 counts at ±2g mode.
Tomas
1,826 次查看
bobpaddock
Senior Contributor III
How well does this work out if the device is moving, in uncontrolled ways,  while self-test is being done?